Johncifelli
Junior
- Joined
- Sep 21, 2020
- Messages
- 23
- Reaction score
- 0
I just purchased some pH strips. Dipped the strip in my wine the strip came up deep burgundy. Which doesn't match any of the colors on the test kit. Not even close. Help!
Do you recommend any specific pH meterThose ph test strips are particularly useless on red wines. unfortunately a meter is probably required. and you really want one that is +/- 0.02
Do you recommend any specific pH meter
I have a Milwaukee MW101 or 102, I forget exactly the number, they are somewhat expensive at about $100, but I believe many of us have one.
Doing a very, very quick google search, I stumbled across two that seem to be very good at a much lower price point:
https://www.amazon.com/Accuracy-Quality-Measurement-Brewing-Laboratory/dp/B08FRFZRYS - $16 - not sure of the accuracy of this one, one place the copy says +/- 0.01 another +/- 0.1
Amazon.com - $32 - this one has better accuracy +/- 0.01
One caution is the calibration procedure on both of these, with the Milwaukee, it is a 2 point calibration and solution is readily available, inexpensively. Both of these do a three point calibration. For wine, since we are usually measuring something down between 3 and 4, I would prefer, if you did a third it was down low, instead of up high, like these have you do, but they seem to be reasonable meters.
Accuracy does matter, since PH is a logarithmic measurement the difference between 3.4 and 3.39 is 10 times.
Accuracy does matter, since PH is a logarithmic measurement the difference between 3.4 and 3.39 is 10 times.
pH is indeed logarithmic, but that greatly overstates the case. The difference between 3 and 4 is 10 times. Your example only represents a few percent difference (2.3%).
But here's the problem, I think. If I am calibrating my pH meter I need to dilute the buffering powders in distilled water. But I would assume that the distilled water is always pH 7.0 . If that distilled water is other than 7.0 how does that affect the buffering solution? and if it DOES affect the solution how does that solution then affect the calibration of the meter? My meter is"recognizing" the nominal values of the buffering solution but is it then compensating for the actual value given the pH of the distilled water being other than 7.0? In other words, how do you confirm the calibration of your pH meter if the only way to confirm this is to use the meter to test a solution that may or may not be the expected value? What I have been doing is using test strips to check whether both they and the meter are giving me essentially the same value (and the strips are good for plus or minus .02 or thereabouts
The whole point of a buffering solution is that its pH will be independent of the water you used to mix it. So, if the water is off from 7.0, it will have no effect on the pH of the solution.
Example is this sentence from the article "What is the pH of Distilled Water?: "Another situation where distilled water may be preferable to use than regular water is in cooling systems for cars. Ions of minerals commonly found within tap water often prove damaging to lead-acid batteries, so distilled water is more likely to be used by cooling systems. "
Exactly - Haven't seen one in a long time.Not to mention, I don't think it is even possible to buy a non-sealed lead-acid battery any more.
aha... That is incredibly useful. Thank you... always happy to learn and in fact I tend to learn something new every single day.The whole point of a buffering solution is that its pH will be independent of the water you used to mix it. So, if the water is off from 7.0, it will have no effect on the pH of the solution.
Enter your email address to join: